Digital SHG vs commercial filters
Posted: Mon Aug 17, 2015 10:52 pm
There has been some interesting discussions on the performance of the commercial filters and what the digital SHG can offer.
http://solarchat.natca.net/viewtopic.php?f=9&t=17095
Let me say a few words about the digital SHG....
A digital spectroheliograph (SHG) is no more than a combination of a telescope with a high dispersion spectrograph and a fast frame mono camera. There are about ten digital SHG active around the world at the present time.
(A spectrohelioscope (SHS) is a visual adaptation of the SHG - See Fred Veio's documents etc. There are only a handful of SHS still in existence and maybe two or three still in active use.)
The spectroheliogram image is produced by obtaining a high resolution spectral image, as an AVI file of the target wavelength as the slit in the SHG scans across the solar surface and then extracting a strip, a few pixels wide from the AVI and combining the strips into a mosaic image - a spectroheliogram.
The final solar surface (spatial) resolution obtained is exactly the same as would be obtained from the same telescope with any commercial filter (white light or narrow band).
However, the filter images obtained with a similar camera can be enhanced by selecting the best quality images, stacking and wavelet processing. This allows the frame rate to freeze (or at least improve) the momentary seeing conditions. The spectroheliogram doesn't (at the moment) have this luxury, each individual strip is an individual image.
The imaging bandwidth used is fixed in the commercial filters (we'll come back to the issue of leakage later) and for narrowband filter usually around 0.7 to 0.3A. With a SHG, any bandwidth can be selected for the imaging process.
The spectral resolution depends on four key factors:
The width of the entrance slit gap
The spectrograph optics (collimator/ imaging lenses)
The grating groove density (l/mm)
and the pixel size of the camera.
To get a final bandwidth comparable with the filters, say 0.3A, the spectrograph spectral resolution needs to be 0.3A or slightly better. This resolution is controlled by the dispersion (A/pixel) and Nyquist sampling. Dispersion (A/pixel) is NOT the same as resolution (A). The 0.3A bandwidth would normally require a dispersion of at least 0.1A/pixel.
This can easily be obtained with a slit gap close to 2-3 times the pixel size - a 5micron pixel needs a 10 to 15micron effective slit combined with a long focal length optical spectrograph system and a large l/mm grating (>1200 l/mm)
It's normal to use a reference lamp to verify resolution. The emission lines (say neon) are measured and the FWHM noted. Also the spectrograph is calibrated in wavelength based on the dispersion. We can then accurately determine the wavelength position of the target central wavelength - it's a very easy task then to nominate say "on band" or 0.2A or 0.7A "off band" to access the red/ blue wings (This is difficult with the filters)
In summary:
- The surface resolution of the SHG is exactly the same as would be achieved with the same telescope in a single frame exposure.
- The imaging bandwidth of the SHG can be accurately selected in width and wavelength.
Any questions/ comments?
http://solarchat.natca.net/viewtopic.php?f=9&t=17095
Let me say a few words about the digital SHG....
A digital spectroheliograph (SHG) is no more than a combination of a telescope with a high dispersion spectrograph and a fast frame mono camera. There are about ten digital SHG active around the world at the present time.
(A spectrohelioscope (SHS) is a visual adaptation of the SHG - See Fred Veio's documents etc. There are only a handful of SHS still in existence and maybe two or three still in active use.)
The spectroheliogram image is produced by obtaining a high resolution spectral image, as an AVI file of the target wavelength as the slit in the SHG scans across the solar surface and then extracting a strip, a few pixels wide from the AVI and combining the strips into a mosaic image - a spectroheliogram.
The final solar surface (spatial) resolution obtained is exactly the same as would be obtained from the same telescope with any commercial filter (white light or narrow band).
However, the filter images obtained with a similar camera can be enhanced by selecting the best quality images, stacking and wavelet processing. This allows the frame rate to freeze (or at least improve) the momentary seeing conditions. The spectroheliogram doesn't (at the moment) have this luxury, each individual strip is an individual image.
The imaging bandwidth used is fixed in the commercial filters (we'll come back to the issue of leakage later) and for narrowband filter usually around 0.7 to 0.3A. With a SHG, any bandwidth can be selected for the imaging process.
The spectral resolution depends on four key factors:
The width of the entrance slit gap
The spectrograph optics (collimator/ imaging lenses)
The grating groove density (l/mm)
and the pixel size of the camera.
To get a final bandwidth comparable with the filters, say 0.3A, the spectrograph spectral resolution needs to be 0.3A or slightly better. This resolution is controlled by the dispersion (A/pixel) and Nyquist sampling. Dispersion (A/pixel) is NOT the same as resolution (A). The 0.3A bandwidth would normally require a dispersion of at least 0.1A/pixel.
This can easily be obtained with a slit gap close to 2-3 times the pixel size - a 5micron pixel needs a 10 to 15micron effective slit combined with a long focal length optical spectrograph system and a large l/mm grating (>1200 l/mm)
It's normal to use a reference lamp to verify resolution. The emission lines (say neon) are measured and the FWHM noted. Also the spectrograph is calibrated in wavelength based on the dispersion. We can then accurately determine the wavelength position of the target central wavelength - it's a very easy task then to nominate say "on band" or 0.2A or 0.7A "off band" to access the red/ blue wings (This is difficult with the filters)
In summary:
- The surface resolution of the SHG is exactly the same as would be achieved with the same telescope in a single frame exposure.
- The imaging bandwidth of the SHG can be accurately selected in width and wavelength.
Any questions/ comments?